Further Elaboration: You might ask this question to try and understand the low frequency response to determine if the Geode amplifiers effectively have a flat response from DC up. Does the Geode go down to DC for example?
Answer: We apply anti-alias filters to the data to prevent out of band noise from being introduced into the data.
The filters are Set with a corner frequency at ¾ of the Nyquist frequency (1/2 the sampling frequency) for almost all of the sampling frequencies. When we sample at 1.5625 uS, the Nyquist frequency is 32kHz and the filters are Set at 20 kHz, because of limitations of our electronics.
There are two single-pole filters: one analog and one Digital. The analog filter is a simple RC filter with 1uF +/-5% and 100kOhm +/- 1%. We short out the capacitor in the conversion process. The Digital filter is software controlled when the option is registered, so it can be switched in or out in the field. It is an IIR Butterworth with a -3dB corner at 0.9Hz for 48ksps sampling, and 0.6Hz at all the lower sample rates.
Also, we offer software and hardware options to modify the low end frequency response of the Geode.
Low-end bandwidth modification:
1.5 Hz, P/N 28311-37
0.6 Hz, P/N 28147-01
DC, P/Ns 28147-02, 28311-37per system plus
MFAM-SuperMag (LCS100S) or MFAM-SX (LCS100X) can be configured to achieve dead-zone-free operation, in which the combined sensor is always active no matter where in the world and which Direction the device is oriented.
To Set up the dead-zone-free configuration:
Make sure the MFAM is in the "One Sensor (No Dead Zone) run mode. If not, please refer to "How to switch the operating mode for SuperMag MFAM" on our website found here:
How to switch the operating mode for SuperMag (LCS100S) MFAM?.
Orient two sensors orthogonally. The ideal relative orientation is shown below (also in the test report in the USB drive shipped with the unit). This configuration works for both the "Low Heading Error" and the "Low Noise" modes.
Low Heading Error Only mode
If you only wish to run the sensor in the "Low Heading Error" mode (note that SX MFAM does NOT have the "Low Noise" mode), a simpler configuration, as shown below, can also achieve the dead-zone-free operation.
If you purchase a low-noise MagArrow (I or II), the factory default is the low-noise mode.
You can also look up the current running mode by loading the following website (in Instrument Status) after connecting to the MagArrow:
192.168.1.1/systemtest.cgi
To switch to a Different mode, please download the special software below and follow the instructions in the readme file:
Attachment : SwitchModes.zip
Please note that even in the low-noise mode, the heading effect is typically only ~20% worse than that in the low-heading mode. For most applications, you don't need to switch to the low-heading mode.
You can 3D print a dead-zone-free sensor holder from the attached .step file.
Attachment : Sensor Holder Perm Shell_Thin Wall_v10.zip
To Set the time on a MetalMapperII Panasonic Toughpad tablet:
1) Using the Applications menu, open a terminal window
2) Set the date and time you want, using the following example as a guide:
sudo date -s "2024-02-20 03:25:55 PM PDT"
You may need to enter the sudo password (which is the same as the password for the standard "geometrics" user.
3) To change the timezone, use the terminal window to get a list of timezones:
ls /usr/share/zoneinfo
// To drill into a region, expand the command as follows:
ls /usr/share/zoneinfo/Pacific Set the local timezone using this command:
// Open the timezone file to edit it. You may need to enter your user password
sudo mousepad /etc/timezone
// Type in the timezone name, of the timezone you want to use, on a single line in the file, for example like this:
Pacific/Guam
If you want to see your time in UTC, Set the timezone to "Etc/UTC". If you want to Set the time in UTC, then when you Set the time using the "date" command, use UTC as follows:
sudo date -s "2024-02-20 03:25:55 PM UTC"
The time associated with each data point in a SEG-2 data file generated by a Geode is related to the time of the
“trigger” event which was instrumental in the production of the file and its content.
The Trigger Master and Trigger Distribution
The trigger event occurs at the Geode designated within the Controller software as the Trigger Master.
Although all Geodes are capable of being Trigger Masters, there must be one and only one Trigger Master in
any properly functioning Geode system. The Controller automatically takes care of this requirement when the
designation is made by a user, and when the system is established at the time of Controller start-up based on a
previous designation (or a default Setting in the case of a “new survey”). All other Geodes in the system will
have their Trigger Master circuit Disabled. A trigger event can be initiated by an external electrical pulse
provided to the trigger input connector of the Trigger Master Geode, or by a command sent via Ethernet from
the Controller to the Trigger Master (usually for test purposes), but only when all conditions are satisfied to
allow data recording. There is also a special trigger initiation situation, called “self-triggering” which will not
be Discussed further here.
Upon acceptance of a trigger event, the Trigger Master will Distribute the trigger signal to all Geodes in the
system, itself included, via an RS-485 network that resides within the Digital interconnect cabling. (Proper
termination of this RS-485 network is automatically taken care of by the Controller.) The trigger signal is
propagated through the cabling and Geodes at the nominal speed of 70% of the speed of light, or approximately
2.1x10^8 m/sec. The maximum Distance of successful propagation depends on a number of factors such as the
number of Geodes involved, the noise environment, the quality of the cables, and the acceptable amount of
timing uncertainty for the particular application. Distances approaching or exceeding 1km should be given
careful attention in this regard. In a 3-D Geode system involving LTUs, each LTU, unlike a Geode, will
reconstruct the trigger signal before sending it on, effectively confining the maximum Distance issue to each
sub-network separated by LTUs. The penalty is an additional delay of about 100nS for each LTU in the route.
The External Trigger Circuit
The external trigger input is capacitively coupled, with a 2mS time constant, to the midpoint of a resistive
voltage Divider. The voltage Difference between the two ends of the Divider constitute a voltage "window",
which size is Set by the trigger sensitivity parameter and can range from essentially zero at the highest
sensitivity, to about +/- 2.5V at the lowest sensitivity. The Geode will trigger (if enabled) if and when the
coupled signal exceeds the window, in either Direction (i.e., positive or negative going). The signal, after the
capacitor, is clamped by Diodes to the range between the trigger signal ground and +5VDC.
The trigger detector output is Disabled when the system is Disarmed, during a parameter change, and during a
shot, up to the trigger hold-off time after the end of the shot. The trigger hold-off time is a parameter Set by the
user.
Preceding the coupling capacitor (i.e., essentially the node accessible at pin A of the external connector), there
is a 3.3K-Ohm pull-up resistor to +5VDC (relative to pin B). Also a fast transient suppressor clamps the input
at about +/-14VDC. It is advised that the DC + AC level of any voltage applied to pin A relative to pin B be
kept within the range of +/-7V, giving some margin of safety.
If a DC voltage somewhat less than +5VDC is applied when the connector is first mated, the instrument may
trigger at that moment. But, subsequently, because of the capacitive coupling, it will trigger on the next positive
or negative going pulse that exceeds the window level. If the duration of the applied voltage pulse is less than
the record length + delay time + hold-off time, then the Geode will effectively be ready to trigger on the same
edge of another similar pulse.
Sub-sample Synchronization
The Geode supports a sub-sample timing synchronization feature used for synchronizing the data acquisition
after a trigger event to the Distributed trigger signal, so that subsequent time points will be known to within 1/32
(~1/20 at the fastest two sampling rates) sample interval. It does this by increasing the sample interval at the
trigger time by 0 to 31/32 of a sample interval in increments of 1/32, so that the first sample after the trigger
would represent a time of one sample interval after the trigger event, with a tolerance within 1/32 of a sample
interval. The following samples continue from there at the expected intervals. For example, with a selected
sampling interval of ¼ mS and a recording delay of 0mS, the first sample in the recorded file for each channel
would represent data at 250 to 258uS after the trigger event.
This of course potentially introduces a small Discontinuity at the time of the trigger, observable depending on
the nature of the channel waveform(s). (The zero-phase anti-alias filter will smear the Discontinuity into the
nearby samples both before and after, consistent with the bandwidth of the filter.) Sub-sample synchronization
can be Disabled if it is deemed to be detrimental for the particular application, at the expense of losing the 1/32
interval timing accuracy.
Timing Errors
The principal errors in Geode timing are of two types: those associated with the trigger mechanism and which
are static over the duration of the record, and those associated with the time base and which change over the
duration of the record. Excluding the trigger propagation delay mentioned above, the trigger timing uncertainty
is about 1uS. The known fixed errors have been lumped together and are reported in the SEG-2 file trace
headers as channel SKEW. (The actual channel skew is zero, since all channels are effectively sampled
simultaneously, but the SKEW value in the header is used as the only place permitting small timing corrections.
Note that the SKEW value for every channel is identical.) If the size of this correction is important to the
application, the SKEW value should be added to the calculated time points when the data is being processed.
The Geode time base has a +/-15ppm stability over temperature (-20C to +70C) and component variations.
Thus time drift relative to absolute time and relative to other Geodes is possible. (However, all channels within
any Geode enclosure use the same time base, so there is no relative drift between channels in the same
enclosure.) Therefore timing uncertainty increases from that existing at the time of the trigger until the time of
the next trigger (or end of record).
Special Timing Issues Involved with “Continuous” Recording
“Continuous” recording is a method that allows unending 100% time coverage with recorded Geode data. It
produces a series of time-overlapped records created by the use of a negative time delay Set equal to the record
length such that each record consists of completed history at the time of the trigger event. This technique
circumvents the problem of data transmission overrunning data acquisition. The principle constraint is that the
cycle time from trigger to trigger must always be less than the chosen record length. Otherwise, gaps rather
than overlap would result. Commonly it is used with GPSderived
triggering in order to provide time-stamping of each trigger event.
Upon consideration of the above, it will become clear that the time-stamp associated with a particular trigger
event will pertain to the data in the following record, not to the data in the record in which the time-stamp is
written. This comes about because the trigger event ends the record.
Because there is data overlap between records, the precise trigger point in the following record at which the
time-stamp applies can be found by comparison of the data values at the end of the former record with those
near the beginning of the subsequent record. The overlapping data will be exactly identical in both records
(since they are read from the same memory location, twice). The earliest data in the subsequent record that
goes beyond the data of the previous record is the data that is one sample interval (assuming sub-sample
synchronization is enabled) past the time-stamp.
Note well that this comparison must be made independently for at least one channel of each 8-channel Geode
board Set, because the Discrete time at which data values are written to the memory buffer, relative to the trigger
event, is a function of each individual board Set in the Geode system.
Correct GPS Time-Stamping
There are Differences between various GPS models that can affect accurate time stamping. The 1PPS signal
from a GPS has a “timing edge” and return edge, of which only the former is the true whole-second edge.
Some models use a rising edge as the timing edge, some the falling edge, and some have it selectable. Consult
the GPS manual to determine the definition of its timing edge. As indicated earlier, the Geode can be triggered
on either a rising or falling edge. It is important to insure that the Geode is being triggered on the proper edge
in order to avoid timing that may be a fraction of a second off. This is expanded upon below.
Some GPS units provide a very narrow timing pulse, others one that has a nearly 50/50 duty cycle. For the
narrow pulse units, almost certainly it is the leading edge (rising or falling) that is the “timing edge”. This case
can be easily handled by using the Geode Trigger Hold-off feature. If a 10-second cycle time is desired, Set the
Trigger Hold-off time to about 9.5 seconds. In this case, there is a very small chance that the very first trigger
could occur on the wrong (trailing) edge, but from then on the leading edge will be used as the triggering edge.
If the GPS provides a 50/50 duty cycle edge, and it is not alterable, then the Geode by itself could as easily start
on the wrong edge as on the correct timing edge, and continue thusly until restarted. For this case, Geometrics
can provide a Trigger Timing Interface Box (TTIB) that will correct the situation. The TTIB can be
programmed to respond only to the correct edge (rising or falling), change the polarity if needed, and gate
through only one of every N 1PPS pulses, where N is programmable. (The TTIB also incorporates an alarm
system that can provide a remote alert if a record is missed.)
Another potential issue comes from the variations between GPS models of the time that the serial time string
(containing the time value of the associated 1PPS) is issued relative to the 1PPS itself. The Geode Controller
attempts to pick the correct serial string based on a calculation involving the known record length, the PC times,
and the trigger notification message from the Geodes. But if the GPS issues the serial string at an unusual time
(and the time has been seen to vary somewhat with a given GPS unit) then it could pick up the incorrect time,
off by 1 second. If rare, it can be subsequently detected and corrected during data processing, but if consistent
it may not be easily detected. Again, the TTIB can accommodate the situation by only gating through to the Controller PC the string belonging to the gated-through 1PPS pulse. The Controller Serial Input Time Window can then safely be widened to 2 seconds (assuming the
cycle time is more than 2 seconds) if need be, to expand the Controller’s search for the string around the calculated trigger time.
A seismograph with an active trigger input like the Geode Seismograph or ES-3000 Seismograph can be triggered many Different ways. The most commonly used methods are with a trigger switch or a trigger geophone. Typically a trigger switch (known as a hammer switch) is attached to the handle of a sledgehammer near the striking end, so when the sledgehammer is hit against a striker plate to create an active source of energy, the piezoelectric crystal in the hammer switch is activated and the seismograph is triggered to record data along the preset parameters. A trigger geophone does this too, but it is placed near the source itself, and is more commonly used with larger energy sources like a propelled energy generator.
If the seismograph isn't triggering with either a hammer switch or a trigger geophone, then the signal may be weak, so turning up the sensitivity could be a workable solution. If the sensitivity is Set too high in SCS then false triggers might be encountered. In most situations having the sensitivity Set to the middle works best.
Depending on where the trigger geophone is it, there may be a Difference between when it is triggered and when a hammer switch would have triggered. Especially in soft ground the trigger geophone signal may be delayed. In general the hammer trigger is a more reliable timing device.
The Differences in trigger time when using a trigger geophone could be due to things like not striking the center of the plate or Differences in the strength of the impact.
More trigger circuit information:
The seismograph can be triggered by shorting the two input pins A and B on the trigger connector of the seismograph. In fact, that is what the hammer switch does (contact closure device) when it impacts a striker plate. The inertia of the impact causes a momentary closure in the device, which in turn, triggers the Geode. There are no internal components that need to be added. Externally, you could construct trigger device or switch, if that is what you desire.
If you were to measure the pins on the Trigger connector on your seismograph (pin A +, pin B -) you would see about 5VDC. The trigger circuit will sense a contact closure or a pulse.
The Geode trigger input is capacitively coupled, with a 2mS time constant, to the midpoint of a resistive voltage Divider. The voltage Difference between the two ends of the Divider constitute a voltage "window", which size is Set by the trigger sensitivity parameter and can range from essentially zero at the highest sensitivity, to about +/- 2.5V at the lowest sensitivity. The Geode triggers (if enabled) if and when the coupled signal exceeds the window, in either Direction. The signal, after the capacitor, is clamped by Diodes to the range between the trigger signal ground and +5VDC.
The trigger detector output is Disabled when the system is Disarmed, during a parameter change, and during a shot, up to the trigger hold-off time after the end of the shot. The trigger hold-off time is a parameter Set by the user.
Preceding the coupling capacitor (i.e., essentially the node accessible at pin A of the external connector), there is a 3.3K-Ohm pull-up resistor to +5VDC (relative to pin B). Also a fast transient suppressor clamps the input at about +/-14VDC. It is advised that the DC + AC level of any voltage applied to pin A relative to pin B be kept within the range of +/-7V, giving some margin of safety.
If a DC voltage somewhat less than +5VDC is applied when the connector is first mated, the instrument may trigger at that moment. But, subsequently, because of the capacitive coupling, it will trigger on the next positive or negative going pulse that exceeds the window level. If the duration of the applied voltage pulse is less than the record length + delay time + hold-off time, then the Geode will effectively be ready to trigger on the same edge of another similar pulse.
The subject of "Bandwidth" comes up often when Discussing cesium magnetometers. There are two Different aspects of bandwidth that are Different and need to be Differentiated:
The cesium magnetometer uses an atomic resonance of the Cs 133 atom (see note 1 below) which varies proportional to the ambient magnetic field. This atomic resonance is used to Set/control the frequency of an oscillator. Therefore the output signal from the magnetometer is a *frequency* which is proportional to the earth's magnetic field at a coefficient of 3.498572 Hertz per nT. Thus the output frequency (called the Larmor frequency) varies from roughly 70KHz at the equator to 350 KHz at the poles.
Because the cesium magnetometer is an oscillator, and because phase is important in an oscillator, the "Bandwidth" of the electronics in the magnetometer must be at least 10 times higher than the maximum output frequency of 350 Khz, or roughly 3.5 MHz.
This bandwidth should not be confused with the magnetic field measurement "Bandwidth", or how fast of a magnetic field change can be measured. To put a scaler value on any magnetic field reading the output frequency of the magnetometer must be counted, and then scaled appropriately to get a field reading in nanoTeslas. The counting process involves opening a gate period, counting the number of Larmor (frequency) cycles that occur, Divide that number by the precise time interval of the gate period. then scale that value by Dividing by the 3.498572 Hz / Larmor coefficient. You get one reading per gate period, which by default is five or ten hertz (200mS to 100 mS gate period). What you get for a reading during any gate period is the time interval average of the Larmor frequency over that period.
The transfer function of a "time interval averaged" signal is [sine(x) / x] with the first zero falling at the sample frequency. Thus if the G-882 is sampling at 10 hertz the maximum resolvable magnetic field change is roughly 5 hertz.
The sample interval of the G-882 is adjustable by sending commands to it. If the sample rate is Set to 20 hertz the measurement bandwidth will double (from a 10 hertz sample rate) but the base line noise will go up as well.
It should also be noted that the basic system noise level of the G-882 for a stationary sensor is Set by the counter resolution - not by the signal to noise ratio of the oscillator electronics. If the sensor is tilted away from its optimum orientation the magnetometer signal will decrease (and therefore the signal to noise ratio), but the counted field output will not show any significant degradation until the sensor is approaching the dead zone (where the signal is really low).
It is a fairly simple task to connect a GPS to a G-858 magnetometer. You can use the External I/O cable assembly and a null modem to connect the G-858 to most GPS receivers.
Null modems are available from Geometrics or from local Radio Shack or computer stores. The Null modem pin configuration for GPS receivers that have a data cable compatible with 9 pin IBM PC COM ports has male pins on both sides.
The GPS should be Set to output NMEA data that contains the $GPGGA sentence. Be sure to Set the RS232 protocol to 9600 Baud, 8 Databits, 1 Stop Bit and No Parity.
The G-858 must have its serial port Set to the same baud rate as the GPS.
You can use System Setup -> Com & Field Note String Setup -> Chat Mode to determine whether correct communications have been established.
The subject of "Bandwidth" comes up often when Discussing cesium magnetometers. There are two Different aspects of bandwidth that are Different and need to be Differentiated:
The cesium magnetometer uses an atomic resonance of the Cs 133 atom (see note 1 below) which varies proportional to the ambient magnetic field. This atomic resonance is used to Set/control the frequency of an oscillator. Therefore the output signal from the magnetometer is a *frequency* which is proportional to the earth's magnetic field at a coefficient of 3.498572 Hertz per nT. Thus the output frequency (called the Larmor frequency) varies from roughly 70KHz at the equator to 350 KHz at the poles.
Because the cesium magnetometer is an oscillator, and because phase is important in an oscillator, the "Bandwidth" of the electronics in the magnetometer must be at least 10 times higher than the maximum output frequency of 350 Khz, or roughly 3.5 MHz.
This bandwidth should not be confused with the magnetic field measurement "Bandwidth", or how fast of a magnetic field change can be measured. To put a scaler value on any magnetic field reading the output frequency of the magnetometer must be counted, and then scaled appropriately to get a field reading in nanoTeslas. The counting process involves opening a gate period, counting the number of Larmor (frequency) cycles that occur, Divide that number by the precise time interval of the gate period. then scale that value by Dividing by the 3.498572 Hz / Larmor coefficient. You get one reading per gate period, which by default is five or ten hertz (200mS to 100 mS gate period). What you get for a reading during any gate period is the time interval average of the Larmor frequency over that period.
The transfer function of a "time interval averaged" signal is [sine(x) / x] with the first zero falling at the sample frequency. Thus if the G-882 is sampling at 10 hertz the maximum resolvable magnetic field change is roughly 5 hertz.
The sample interval of the G-882 is adjustable by sending commands to it. If the sample rate is Set to 20 hertz the measurement bandwidth will double (from a 10 hertz sample rate) but the base line noise will go up as well.
It should also be noted that the basic system noise level of the G-882 for a stationary sensor is Set by the counter resolution - not by the signal to noise ratio of the oscillator electronics. If the sensor is tilted away from its optimum orientation the magnetometer signal will decrease (and therefore the signal to noise ratio), but the counted field output will not show any significant degradation until the sensor is approaching the dead zone (where the signal is really low).
Hi All,
We are using the Teensy 4.1 as a logger (Adafruit GPS is linked with the Teensy) for the MFAM SX. Below is the code for the Arduino IDE (Teensy), for people who might find it useful,
Roi
#include <NativeEthernet.h>
#include <SD.h>
// ---- NETWORK CONFIGURATION ----
byte mac[] = { 0xDE, 0xAD, 0xBE, 0xEF, 0xFE, 0xED };
IPAddress ip(192, 168, 2, 10);
IPAddress mfamIP(192, 168, 2, 2);
uint16_t mfamPort = 1000;
EthernetClient client;
// ---- PACKET STRUCTURE ----
const int PACKET_SIZE = 1380;
const int SAMPLE_SIZE = 32;
const int HEADER_SIZE = 16;
const int NUM_SAMPLES = 40;
const int SD_CHIP_SELECT = BUILTIN_SDCARD;
uint8_t buffer[PACKET_SIZE];
int bufferPos = 0;
// ---- SD CARD ----
File logFile;
bool sdReady = false;
unsigned long sampleCount = 0;
unsigned long fileStartTime = 0;
char filename[32];
// ---- AUXILIARY CHANNEL STORAGE ----
double gyroX = 0, gyroY = 0, gyroZ = 0, gyroT = 0;
double accelX = 0, accelY = 0, accelZ = 0, accelT = 0;
double compassX = 0, compassY = 0, compassZ = 0, compassT = 0;
// ---- GPS FROM ADAFRUIT MODULE ON SERIAL1 (Pin 0 = RX) ----
char gpsBuffer[256];
int gpsBufferPos = 0;
char gpsString[128] = "";
char gpsDate[12] = "00/00/00";
char gpsTime[16] = "00:00:00.000";
bool gpsFix = false;
uint8_t tsStatus = 0;
// ---- OUTPUT CONTROL ----
// Set to 1 to log every sample, 10 for 100Hz, 20 for 50Hz, etc.
const int DOWNSAMPLE_FACTOR = 1; // 50 Hz output
int downsampleCounter = 0;
// How many minutes per file. Set to 10, 20, 60, etc.
const int FILE_MINUTES = 10;
// ---- LED INDICATOR ----
// Off = starting up
// Very slow blink (every 3 sec) = connected and logging
// Fast blink (4/sec) = connected but no data arriving
// Solid on = no SD card
// 3 quick flashes then pause = cannot connect to MFAM
const int LED_PIN = 13;
unsigned long lastBlinkTime = 0;
bool ledState = false;
unsigned long lastDataTime = 0;
// ---- FUNCTION PROTOTYPES ----
void createNewFile();
void parsePacket(uint8_t* pkt);
void parseAuxChannels(uint8_t* sample, uint16_t frameID);
void readGPS();
void parseGPRMC(char* sentence);
void writeSample(unsigned long timestamp, uint16_t fiducial, double mag1, uint16_t mag1s,
double mag2, uint16_t mag2s, uint16_t sysStatus);
int16_t toSigned16(uint16_t val);
void createNewFile() {
static int fileNumber = 0;
// On first call, find the next available file number
if (fileNumber == 0) {
char testName[32];
for (int i = 1; i <= 99999; i++) {
snprintf(testName, sizeof(testName), "MFAM_%05d.txt", i);
if (!SD.exists(testName)) {
fileNumber = i - 1; // Will be incremented below
break;
}
}
}
fileNumber++;
snprintf(filename, sizeof(filename), "MFAM_%05d.txt", fileNumber);
logFile = SD.open(filename, FILE_WRITE);
if (logFile) {
logFile.println("Mag 1,Mag 2,Fid,SysS,Mg1S,Mg2S,Gyro X,Gyro Y,Gyro Z,Gyro T,Accel X,Accel Y,Accel Z,Accel T,CompassX,CompassY,CompassZ,Comp T,Date,Time,TS Status,GPS");
logFile.flush();
fileStartTime = millis();
Serial.print("Logging to: ");
Serial.println(filename);
} else {
Serial.print("ERROR: Could not create ");
Serial.println(filename);
}
}
void Setup() {
Serial.begin(115200);
delay(2000);
// Start GPS serial port (Adafruit Ultimate GPS defaults to 9600 baud)
Serial1.begin(9600);
pinMode(LED_PIN, OUTPUT); DigitalWrite(LED_PIN, LOW);
memset(gpsString, 0, sizeof(gpsString));
// Initialize SD card
if (SD.begin(SD_CHIP_SELECT)) {
sdReady = true;
Serial.println("SD card ready.");
} else {
Serial.println("WARNING: No SD card found. Serial output only."); DigitalWrite(LED_PIN, HIGH);
}
// Initialize Ethernet
Ethernet.begin(mac, ip);
if (Ethernet.hardwareStatus() == EthernetNoHardware) {
Serial.println("ERROR: No Ethernet hardware found!");
while (true) {}
}
Serial.print("Teensy IP: ");
Serial.println(Ethernet.localIP());
Serial.print("Connecting to MFAM at ");
Serial.print(mfamIP);
Serial.print(":");
Serial.println(mfamPort);
if (client.connect(mfamIP, mfamPort)) {
Serial.println("Connected to MFAM!");
} else {
Serial.println("Connection failed!");
}
Serial.println("Waiting for GPS fix...");
if (sdReady) {
createNewFile();
}
Serial.println("Mag 1,Mag 2,Fid,SysS,Mg1S,Mg2S,Gyro X,Gyro Y,Gyro Z,Gyro T,Accel X,Accel Y,Accel Z,Accel T,CompassX,CompassY,CompassZ,Comp T,Date,Time,TS Status,GPS");
}
void loop() {
// Always read GPS data from Serial1
readGPS();
if (!client.connected()) {
Serial.println("Disconnected. Reconnecting...");
if (sdReady && logFile) {
logFile.flush();
}
for (int i = 0; i < 3; i++) { DigitalWrite(LED_PIN, HIGH);
delay(100); DigitalWrite(LED_PIN, LOW);
delay(100);
}
delay(1400);
client.connect(mfamIP, mfamPort);
if (client.connected()) {
lastDataTime = millis();
}
return;
}
while (client.available()) {
buffer[bufferPos] = client.read();
bufferPos++;
if (bufferPos >= PACKET_SIZE) {
parsePacket(buffer);
bufferPos = 0;
lastDataTime = millis();
}
}
// LED patterns
if (sdReady) {
if (millis() - lastDataTime > 3000) {
if (millis() - lastBlinkTime > 125) {
ledState = !ledState; DigitalWrite(LED_PIN, ledState ? HIGH : LOW);
lastBlinkTime = millis();
}
} else {
if (millis() - lastBlinkTime > 1500) {
ledState = !ledState; DigitalWrite(LED_PIN, ledState ? HIGH : LOW);
lastBlinkTime = millis();
}
}
}
// New file every FILE_MINUTES minutes
if (sdReady && logFile && (millis() - fileStartTime > (unsigned long)FILE_MINUTES * 60UL * 1000UL)) {
logFile.close();
createNewFile();
}
}
// ---- GPS READING FROM ADAFRUIT MODULE ON SERIAL1 ----
void readGPS() {
while (Serial1.available()) {
char c = Serial1.read();
if (c == '$') {
gpsBufferPos = 0;
}
if (gpsBufferPos < (int)sizeof(gpsBuffer) - 1) {
gpsBuffer[gpsBufferPos] = c;
gpsBufferPos++;
}
if (c == '\n' || c == '\r') {
gpsBuffer[gpsBufferPos] = '\0';
if (strncmp(gpsBuffer, "$GPRMC", 6) == 0 || strncmp(gpsBuffer, "$GNRMC", 6) == 0) {
// Save full sentence for logging
strncpy(gpsString, gpsBuffer, sizeof(gpsString) - 1);
gpsString[sizeof(gpsString) - 1] = '\0';
// Remove trailing newline/carriage return
int slen = strlen(gpsString);
while (slen > 0 && (gpsString[slen - 1] == '\n' || gpsString[slen - 1] == '\r')) {
gpsString[slen - 1] = '\0';
slen--;
}
parseGPRMC(gpsBuffer);
}
gpsBufferPos = 0;
}
}
}
void parseGPRMC(char* sentence) {
// $GPRMC,HHMMSS.sss,A,lat,N,lon,W,speed,course,DDMMYY,...
char copy[256];
strncpy(copy, sentence, sizeof(copy) - 1);
copy[sizeof(copy) - 1] = '\0';
char* token = strtok(copy, ",");
int field = 0;
while (token != NULL && field < 10) {
switch (field) {
case 1: // Time
if (strlen(token) >= 6) {
snprintf(gpsTime, sizeof(gpsTime), "%c%c:%c%c:%s",
token[0], token[1], token[2], token[3], token + 4);
}
break;
case 2: // Fix status
gpsFix = (token[0] == 'A');
break;
case 9: // Date
if (strlen(token) >= 6) {
snprintf(gpsDate, sizeof(gpsDate), "%c%c/%c%c/%c%c",
token[0], token[1], token[2], token[3], token[4], token[5]);
}
break;
}
token = strtok(NULL, ",");
field++;
}
// Update GPS bits of tsStatus
tsStatus = (tsStatus & 0x0C); // Keep MFAM PPS bits (3,2)
tsStatus |= 0x01; // Bit 0: RMC sentence received
if (gpsFix) {
tsStatus |= 0x02; // Bit 1: GPS fix valid
}
}
// ---- MFAM DATA PARSING ----
int16_t toSigned16(uint16_t val) {
if (val > 32767) return (int16_t)(val - 65536);
return (int16_t)val;
}
void parseAuxChannels(uint8_t* sample, uint16_t frameID) {
uint8_t auxID = (frameID >> 11) & 0x07;
uint16_t aux0 = sample[16] | (sample[17] << 8);
uint16_t aux1 = sample[18] | (sample[19] << 8);
uint16_t aux2 = sample[20] | (sample[21] << 8);
uint16_t aux3 = sample[22] | (sample[23] << 8);
switch (auxID) {
case 1:
compassX = toSigned16(aux0) / 0.01333333;
compassY = toSigned16(aux1) / 0.01333333;
compassZ = toSigned16(aux2) / 0.01333333;
compassT = toSigned16(aux3) / 128.0 + 25.0;
break;
case 2:
gyroX = toSigned16(aux0) / 16.384;
gyroY = toSigned16(aux1) / 16.384;
gyroZ = toSigned16(aux2) / 16.384;
gyroT = toSigned16(aux3) / 512.0 + 23.0;
break;
case 4:
accelX = toSigned16(aux0) / 16384.0;
accelY = toSigned16(aux1) / 16384.0;
accelZ = toSigned16(aux2) / 16384.0;
accelT = toSigned16(aux3) / 512.0 + 23.0;
break;
default:
break;
}
}
void writeSample(unsigned long timestamp, uint16_t fiducial, double mag1, uint16_t mag1s,
double mag2, uint16_t mag2s, uint16_t sysStatus) {
char tsStr[12];
snprintf(tsStr, sizeof(tsStr), "%d%d%d%d%d%d%d%d",
(tsStatus >> 7) & 1, (tsStatus >> 6) & 1, (tsStatus >> 5) & 1, (tsStatus >> 4) & 1,
(tsStatus >> 3) & 1, (tsStatus >> 2) & 1, (tsStatus >> 1) & 1, tsStatus & 1);
char sysHex[8], m1sHex[8], m2sHex[8];
snprintf(sysHex, sizeof(sysHex), "%04X", sysStatus);
snprintf(m1sHex, sizeof(m1sHex), "%04X", mag1s);
snprintf(m2sHex, sizeof(m2sHex), "%04X", mag2s);
char line[512];
snprintf(line, sizeof(line),
"%10.4f,%10.4f,%4d,%s,%s,%s,%10.2f,%10.2f,%10.2f,%5.1f,%8.5f,%8.5f,%8.5f,%5.1f,%8.1f,%8.1f,%8.1f,%5.1f,%s,%s,%s,%s",
mag1, mag2, fiducial, sysHex, m1sHex, m2sHex,
gyroX, gyroY, gyroZ, gyroT,
accelX, accelY, accelZ, accelT,
compassX, compassY, compassZ, compassT,
gpsDate, gpsTime, tsStr, gpsString);
if (sdReady && logFile) {
logFile.println(line);
if (sampleCount % 1000 == 0) {
logFile.flush();
}
}
if (sampleCount % 500 == 0) {
Serial.println(line);
}
}
void parsePacket(uint8_t* pkt) {
// Get PPS bits from MFAM system status
uint16_t firstSysStatus = pkt[HEADER_SIZE + 2] | (pkt[HEADER_SIZE + 3] << 8);
uint8_t mfamPPSbits = (firstSysStatus >> 12) & 0x0C;
tsStatus = (tsStatus & 0x03) | mfamPPSbits;
for (int i = 0; i < NUM_SAMPLES; i++) {
int offset = HEADER_SIZE + (i * SAMPLE_SIZE);
uint16_t frameID = pkt[offset] | (pkt[offset + 1] << 8);
uint16_t fiducial = frameID & 0x07FF;
uint16_t sysStatus = pkt[offset + 2] | (pkt[offset + 3] << 8);
uint32_t mag1raw = pkt[offset + 4] | (pkt[offset + 5] << 8) |
((uint32_t)pkt[offset + 6] << 16) | ((uint32_t)pkt[offset + 7] << 24);
double mag1 = mag1raw * 0.05 / 1000.0;
uint16_t mag1status = pkt[offset + 8] | (pkt[offset + 9] << 8);
uint32_t mag2raw = pkt[offset + 10] | (pkt[offset + 11] << 8) |
((uint32_t)pkt[offset + 12] << 16) | ((uint32_t)pkt[offset + 13] << 24);
double mag2 = mag2raw * 0.05 / 1000.0;
uint16_t mag2status = pkt[offset + 14] | (pkt[offset + 15] << 8);
parseAuxChannels(pkt + offset, frameID);
sampleCount++;
downsampleCounter++;
if (downsampleCounter >= DOWNSAMPLE_FACTOR) {
downsampleCounter = 0;
writeSample(millis(), fiducial, mag1, mag1status, mag2, mag2status, sysStatus);
}
}
}
Verify BAUD Rate selected on Dip switches 6,7,8
Note: according to a customer this is how the switches are Set as they are attempting a file transfer at 115200 BAUD.
Connect the G-857 to the computer using the download cable and a properly installed USB/RS-232 adaptor. (Must have FTDI):
Open the MagMap Software.
Select Import>G-857/ASCII
1. Enter known Serial Port (My computer is COM4)
2. BAUD 115200 (Switches 6, 7 and 8 off)
3. Set a good location for the file to written to.
4. Leave Download only, open later unchecked.
5. Leave download time out 2.
Select OK.
On the G-857 press OUTPUT, ENTER when this window opens.
BYTES DOWNLOADED will begin to increment.
Select OK in the window below:
The downloaded Mag file should be Displayed:
If the download sequence does not happen this way, it is usually a problem with:
1. USB/RS-232 Adaptor not FTDI, or driver not installed.
2. COM PORT assignment incorrect, and/or file destination not valid.
3. BAUD RATE mismatch.
4. Faulty I/O Cable PN 16492-01.
5. Faulty Computer.
6. Faulty G-857 console.
Check to make sure that the G-857 is not in Legacy Mode:
To check if Legacy Mode is enabled press:
AUTO-OUTPUT
To turn it off press:
AUTO-OUTPUT-CLEAR
To Turn on:
AUTO-OUTPUT-ENTER
If Legacy Mode was on then the recorded data recorded needs to be downloaded as a G-856:
What Affects Geode Trigger Cycle Times?
If you're trying to optimize your Geode system for faster trigger cycles—especially in high-repeat environments—there are a few key factors to consider. The goal is to ensure that the system completes its entire cycle (trigger → recording → data transfer → re-arming) before the next expected trigger. Here’s what influences that cycle:
🧠 Core Factors That Affect Cycle Times
1. File Size (Sampling Parameters)
Your sample interval and record length Directly affect the size of each data file.You can view the resulting file size in the Acquisition Parameters menu.Larger files take longer to transfer, which delays the re-arm process.
2. Data Transfer Rate
The Geode typically transfers data at around 450–465 kb/sec.Reducing file size is the best way to reduce transfer time and speed up the cycle.
3. Calibration Frequency
By default, the system may attempt to calibrate every N shots, which takes additional time.Go to Options > Calibration and Set "calibrate every N shots" to a large number (e.g., 100000) to prevent unnecessary delays.
4. Recording Delay and Record Length
If you're operating in a region with a consistently deep seafloor, you can add a recording delay and reduce record length accordingly.Example: If the water column is always >0.3s, you can apply a delay of 0.2s and reduce record length by the same amount.This trims your file size and speeds up the transfer/re-arm process.
⚙️ Best Practices
Use the Auto-Trigger function or Set trigger sensitivity to the maximum value for testing.Monitor the cycle timing and adjust acquisition parameters to stay within your trigger window.It's often an iterative process to find the ideal configuration for your environment.
I have Set up a G-882 system here at Geometrics and am receiving data and sending commands using TeraTerm (any terminal emulation program should work). When in normal use mode the Digital add on board sits in front of the G-882 and parses and acts on all commands coming in. There are two versions of the Digital Add On board, which are the GP120 and the GP140. The GP140 is a newer version of the Digital Add ON board. It is the GP140 Digital board that outputs all S/N (and other) information.
I first Set up with the GP140 board (the newer version). I find that the ""RESET" command does work - i.e. it goes into BYPASS mode for a couple seconds, then output the S/N and configuration information, and reverts to normal operation with the Digital depth and altimeter information. But it only works every other time I send it. The first time nothing happens. Then I send it again and it works. This appears to be a bug in the GP140. For some commands the first command after power up or reset are ignored. The second time (and subsequent commands) are executed. The work around seems to be sending the RESET command twice.
I also tried an older G-882 with the GP120 Digital board. The Reset (and other commands worked first time and every time.
BTW, the Digital Board version is in the second line with the S/N information that is sent on power up or Reset.
Some questions:
1) My configuration is one G-882 connected to a PC through the white junction box. Is this your configuration, or do you have concatenated G-882's?
2) Can you get the G-882 to accept any commands (like going into Bypass Mode)? I'm wondering if there is a open link in the command line from the PC to the Digital board.
General information
The most common symptoms of intermittent connection issues are shown below: D_CY shows a background decay signal is either much higher than normal, and/or C_CY is a flat line (doesn’t decay).
Figure 1 Intermittent connection issue.
Where is the failure occurring?
The two most common places where intermittent issues occur are at the two ends of the Rx cable: the joint between the Rx cable and the Cart and the joint between the Rx cable and the EDA box (orange box).
Now we need to identify which joint has the intermittent issue. Set up the MM2x2 in DAM mode.
Collect DAM data while keeping the Cart stationary but tapping one joint.
Collect another DAM data while tapping the other joint.
Analyze the DAM data by plotting the “Monostatic_5” for all 12 Rx channels in Geosoft. The channels having intermittent issues will appear much noisier.
If you have MatLab software, you can download the MatLab code to analyze the DAM data. Example plots are shown below. It is obvious that “ZA” channel has the intermittent issue in Figure 2 and “XB” channel is open in Figure 3 (very flat line, no noise at all). Click here to download the code:
Attachment : Intermittent_noise_full.zip
.
If both DAM and IVS data have the same problematic channel(s), we are confident that the intermittent issues observed in IVS data are repeated in DAM data, and by tapping at that location, we are able to identify the intermittent joint.
Figure 2 Intermittent "ZA" channel.
Figure 3 Open "XB" channel.
What to do next? Disconnect the problematic joint and clean the connectors on both sides thoroughly (using an acid brush and a can of compressed air). Reconnect and try the tapping method again. If the problem goes away (no more noisy channels), the intermittent issue is likely caused by dust.
If cleaning doesn’t fix the problem, swap out the Rx cable and repeat the tapping method. If the problem goes away, it is likely caused by a bad Rx cable.
If there is another Set of EDA and Cart available, swap out the EDA and the Cart to identify the problematic part.
If not, use the tapping location to identify the problematic part.
Fill out the RMA form at .
If it is the Cart, send in the whole system for inspection/repair. You can contact Geometrics for MM2x2 rental if you need to continue your work during the down time.
If it is the EDA, we recommend sending in the EDA only. It will save your repair time since it is much faster to unpack/pack/ship the EDA than the whole system. You can contact Geometrics for EDA rental if you need to continue your work during the down time.
Warning
Please note that this tapping method should ONLY be tried when intermittent issues have been observed in IVS tests. It is NOT recommended to use it as a daily QC test because it does put extra stress on connectors and likely leads to a shortened connector lifetime if applied too often.
Page 1 / 3