Question What is the Detection Range of a Magnetometer?

1 Posts
1 Users
0 Reactions
652 Views
(@gschmauder)
Member Admin
Joined: 6 years ago
Posts: 117
Topic starter  

This is a very common question and for the best answer, please read page 45 of the Applications Manual for Portable Magnetometers.

To give an idea, the rule of thumb is that 1 ton of steel will give 1nT at 100 ft. The distortion caused by the steel in the earth’s field falls off as the cube with distance and is linear with mass. Therefore, at 50 ft, 250 lbs will give 1nT, at 25 ft 30 lbs will give 1nT, at 12 ft 4 lbs will give 1nT. Cables and pipelines fall off at somewhat a different rate (inverse square) so can be seen further for a given mass.


   