The subject of "Bandwidth" comes up often when discussing cesium magnetometers. There are two different aspects of bandwidth that are different and need to be differentiated:
The cesium magnetometer uses an atomic resonance of the Cs 133 atom (see note 1 below) which varies proportional to the ambient magnetic field. This atomic resonance is used to set/control the frequency of an oscillator. Therefore the output signal from the magnetometer is a *frequency* which is proportional to the earth's magnetic field at a coefficient of 3.498572 Hertz per nT. Thus the output frequency (called the Larmor frequency) varies from roughly 70KHz at the equator to 350 KHz at the poles.
Because the cesium magnetometer is an oscillator, and because phase is important in an oscillator, the "Bandwidth" of the electronics in the magnetometer must be at least 10 times higher than the maximum output frequency of 350 Khz, or roughly 3.5 MHz.
This bandwidth should not be confused with the magnetic field measurement "Bandwidth", or how fast of a magnetic field change can be measured. To put a scaler value on any magnetic field reading the output frequency of the magnetometer must be counted, and then scaled appropriately to get a field reading in nanoTeslas. The counting process involves opening a gate period, counting the number of Larmor (frequency) cycles that occur, divide that number by the precise time interval of the gate period. then scale that value by dividing by the 3.498572 Hz / Larmor coefficient. You get one reading per gate period, which by default is five or ten hertz (200mS to 100 mS gate period). What you get for a reading during any gate period is the time interval average of the Larmor frequency over that period.
The transfer function of a "time interval averaged" signal is [sine(x) / x] with the first zero falling at the sample frequency. Thus if the G-882 is sampling at 10 hertz the maximum resolvable magnetic field change is roughly 5 hertz.
The sample interval of the G-882 is adjustable by sending commands to it. If the sample rate is set to 20 hertz the measurement bandwidth will double (from a 10 hertz sample rate) but the base line noise will go up as well.
It should also be noted that the basic system noise level of the G-882 for a stationary sensor is set by the counter resolution - not by the signal to noise ratio of the oscillator electronics. If the sensor is tilted away from its optimum orientation the magnetometer signal will decrease (and therefore the signal to noise ratio), but the counted field output will not show any significant degradation until the sensor is approaching the dead zone (where the signal is really low).
Total field magnetometers like the optically pumped cesium magnetometer are passive devices, they do not send out waves or pulses. They measure distortions in the earth’s normally homogenous magnetic field and can sense distortions due to ferrous objects at great distances.
The basic rule of thumb is that one ton (1000 Kg) of steel or iron will give us a 1nT anomaly at 100 ft. or 30m. Since the amount of distortion falls off as the cube with distance (compare a metal detector which falls off as the inverse 6th power!) and is linear with mass, every time we cut the distance in half, we can see 1/8th the mass. Therefore, we can sense 250 lbs. (100kg) at 50 feet (15m), or 30lbs (15kg) at 25 feet (8m), or 4lbs (2kg) at 12 feet (4m).
However this is not the whole story. The factors given above are for induced magnetic fields only. Many targets also have remnant or permanent magnetic effects (meaning they have become magnetized either in production or by the earth’s field) and can therefore have larger anomalies by a factor of 3 or 5 or more. Also many hollow objects like barrels or other tubular structures appear as though they are solid due to self-shielding from the earth’s field, and thus have much larger anomalies than their mass would predict alone. Pipes fall off as the inverse square and are thus detectable at even greater distances. Please see our Applications Manual for Portable Magnetometers for more information.
The subject of "Bandwidth" comes up often when discussing cesium magnetometers. There are two different aspects of bandwidth that are different and need to be differentiated:
The cesium magnetometer uses an atomic resonance of the Cs 133 atom (see note 1 below) which varies proportional to the ambient magnetic field. This atomic resonance is used to set/control the frequency of an oscillator. Therefore the output signal from the magnetometer is a *frequency* which is proportional to the earth's magnetic field at a coefficient of 3.498572 Hertz per nT. Thus the output frequency (called the Larmor frequency) varies from roughly 70KHz at the equator to 350 KHz at the poles.
Because the cesium magnetometer is an oscillator, and because phase is important in an oscillator, the "Bandwidth" of the electronics in the magnetometer must be at least 10 times higher than the maximum output frequency of 350 Khz, or roughly 3.5 MHz.
This bandwidth should not be confused with the magnetic field measurement "Bandwidth", or how fast of a magnetic field change can be measured. To put a scaler value on any magnetic field reading the output frequency of the magnetometer must be counted, and then scaled appropriately to get a field reading in nanoTeslas. The counting process involves opening a gate period, counting the number of Larmor (frequency) cycles that occur, divide that number by the precise time interval of the gate period. then scale that value by dividing by the 3.498572 Hz / Larmor coefficient. You get one reading per gate period, which by default is five or ten hertz (200mS to 100 mS gate period). What you get for a reading during any gate period is the time interval average of the Larmor frequency over that period.
The transfer function of a "time interval averaged" signal is [sine(x) / x] with the first zero falling at the sample frequency. Thus if the G-882 is sampling at 10 hertz the maximum resolvable magnetic field change is roughly 5 hertz.
The sample interval of the G-882 is adjustable by sending commands to it. If the sample rate is set to 20 hertz the measurement bandwidth will double (from a 10 hertz sample rate) but the base line noise will go up as well.
It should also be noted that the basic system noise level of the G-882 for a stationary sensor is set by the counter resolution - not by the signal to noise ratio of the oscillator electronics. If the sensor is tilted away from its optimum orientation the magnetometer signal will decrease (and therefore the signal to noise ratio), but the counted field output will not show any significant degradation until the sensor is approaching the dead zone (where the signal is really low).